Mass Spectrometry
by
Dr. Avinash Upadhyay
What does a mass spectrometer do?
1. It measures mass better than any other technique.
2. It can give information about chemical structures.
What are mass measurements good for?
To identify, verify, and quantitate: metabolites, recombinant proteins, proteins isolated from natural sources, oligonucleotides, drug candidates, peptides, synthetic organic chemicals, polymers
Pharmaceutical analysis
Bioavailability studies
Drug metabolism studies, pharmacokinetics
Characterization of potential drugs
Drug degradation product analysis
Screening of drug candidates
Identifying drug targets
Biomolecule characterization
Proteins and peptides
Oligonucleotides
Environmental analysis
Pesticides on foods
Soil and groundwater contamination
Forensic analysis/clinical
Applications of Mass Spectrometry
Inlet
Ionization
Mass Analyzer
Mass Sorting (filtering)
Ion
Detector
Detection
Ion Source
• Solid
• Liquid
• Vapor
Detect ions
Form ions
(charged molecules)
Sort Ions by Mass (m/z)
1330
1340
1350
100
75
50
25
0
Mass Spectrum
Summary: acquiring a mass spectrum
Inlet
Ion
source
Mass
Analyzer
Detector
Data
System
High Vacuum System
Mass Spectrometer Block Diagram
Inlet
Ion
source
Mass
Analyzer
Detector
Data
System
High Vacuum System
Mass Spectrometer Block Diagram
Turbo molecular pumps
Inlet
Ion
Source
Mass
Analyzer
Detector
Data
System
High Vacuum System
HPLC
Flow injection
Sample plate
Sample Introduction
Inlet
Ion
Source
Mass
Analyzer
Detector
Data
System
High Vacuum System
MALDI
ESI
FAB
LSIMS
EI
CI
Ion Source
Methods for Ionization�
Method of Ionization is selected on the basis of sample
Ionization of volatile compound:
Two methods are commonly used for volatile compounds:
a) Electron ionization (EI)
ABC + e 🡪 [ABC]+* + 2e
Chemical ionization�
A soft ionization method
Reagent gas reagent ion (Z+) M [MZ]+
Examples:
CH4 + e CH4 + + 2e
CH4 + + e CH5+ + CH3
CH5+ + M CH4 + MH+ ( molecular ion, (M + 1)
Little fragmentation; find M+ which is unstable under EI conditions
High voltage applied
to metal sheath (~4 kV)
Sample Inlet Nozzle
(Lower Voltage)
Charged droplets
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
+
MH+
MH3+
MH2+
Pressure = 1 atm�Inner tube diam. = 100 um
Sample in solution
N2
N2 gas
Partial�vacuum
Electrospray ionization:
Ion Sources make ions from sample molecules�(Ions are easier to detect than neutral molecules.)
�Electrospray ionization �
hν
Laser
1. Sample is mixed with matrix (X) and dried on plate.
2. Laser flash ionizes matrix molecules.
3. Sample molecules (M) are ionized by proton transfer: �XH+ + M 🡪 MH+ + X.
MH+
MALDI: Matrix Assisted Laser Desorption Ionization
+/- 20 kV
Grid (0 V)
Sample plate
MatrixMatrix--assisted laser desorption/ionization assisted ionization
Matrix assisted laser desorption/ionization Matrix-
Inlet
Ion
source
Mass
Analyzer
Detector
Data
System
High Vacuum System
Time of flight (TOF)
Quadrupole
Ion Trap
Magnetic Sector
FTMS
Mass Analyzer
Resolution:
Resolution is the ability of a mass spectrometer to distinguish between ions of different mass-to-charge ratios. Therefore, greater resolution corresponds directly to the increased ability to differentiate ions.
Mass analyzers separate ions based on their
mass-to-charge ratio (m/z)
Quadrupole Mass Analyzer
Uses a combination of RF and DC voltages to operate as a mass filter.
Quadruple Mass Analyzer
mass scanning mode
m1
m3
m4
m2
m3
m1
m4
m2
single mass transmission mode
m2
m2
m2
m2
m3
m1
m4
m2
Quadrupoles have variable ion transmission modes
If an ion is accelerated with a voltage, its resulting velocity is a function of its mass/charge ratio.
Time-of-flight (TOF) Mass Analyzer
Time-of-flight (TOF) Mass Analyzer
+
+
+
+
Source
Drift region (flight tube)
detector
V
Ion Trap analyser:
stable orbits within the assembly.