Cooke Evaporator
Startup & Shutdown
Tool Shutdown
Normal and Emergency Shutdown:
- Push red OFF button labeled Control Power on upper left of the panel (you may have to take off the clear plastic cover - just lift it slightly and pull it towards you.)
- Shut off E-beam circuit breaker if it is on
- If PCW has issues, water valves are at back of machine
Tool Startup and Power Bump Recovery
Power Failure and Startup Procedure:
- Push the black On/Reset button labelled Control Power - it’s  left side of the top panel.  You have to remove the clear plastic to turn it on - lift it slightly and pull it towards you. Â

- Walk around to the back of tool and look at temperature gauge on back of cryo.

- If the gauge on the cryo pump reads <25k and the TC1 cryo pressure is below 40mT, then the cryo is okay. Â
- The system is now okay to run.
- Replace the  clear plastic panel
                          Â
- If the gauge on the cryo pump reads >25k, and/or the cryo pump TC Â pressure is >40mTorr then there is condensation on the cryo and you need to initiate Cryo Pump Regen Procedure